Historic Boundaries and Conservation EffortsHave you tried different points for the resection and checking the fixed (control) point from a different setup to see if the 12 mm error is replicated?
Also, how good are your resection results after the service (calibration) compared to before?
@minbarwinkle I sent back the kit forthem to do it properly and got a replacement TS16 back and did the same face 1/face 2 measurements (same checks) on it and got 7mm error in 150m – what I believe not too bad for a 5″ kit. This also means, that my 1″ total station should get at least the same result if not better…
I’ve tried different resections to do the same checks, however I was checking the calibration of the kit with these checks, not where I am positioned on the site with each resections as the retro targets are not 0mm perfect (there is always a few mm error in resections). Luckily the subcontractor’s engineer is a friend of mine so we keep checking each other’s setting out to make sure we are on the same page, usually we are in 5-6mm tops – what I believe is the difference between the total stations/resections, etc.
This also means, that my 1″ total station should get at least the same result if not better…
Same? Yes, probably. Better? Probably not. Generally, the instrument contributes only a small fraction of the error in the instrument – tribrach – tripod – setup – target prism glass – target rod/tripod centering system.
@gk88 I am using a 1″ Geomax Zoom 95 with the X-Pad software and I do reflectorless resections every day from “stickers ” set by myself before the project started. The distances to the targets are 160m to 250m. I always get the residuals less than 1mm for the distance and less than 2″ for angles. Re-staking points done from previous setups always yield “perfect” results , 1 to 3mm and this is done using a Leica 360 prism.
little update… got back my kit again and “surprisingly” much less error I am getting… on 100 meters I have 1mm (!!!!!) error on the face 2 measurement (face 1 was obviously zero). I wish they cared at the first time.